Legacy flatness dimension is displaying "THE ORIGIN" for Reference Id <2> in summary mode
<< Wade Burton -- 02/27/07 10:12:55>>
Legacy flatness dimension is displaying "THE ORIGIN" for Reference Id <2> in summary mode in the edit window. The flatness dimension shouldn't relate to any reference frame, doesn't reference a second feature because it is a form dimension. No other form dimensions allow a second Reference Id in the summary mode. Plus, it appears this field isn't even used. When I type in feature ids or variables, they don't even appear to accept the text, as the word remains in lower case letters, unlike what happens when any other reference id is entered in.
I discussed this with Don and Bret through e-mail, and Don thinks this should be removed as well.
This was found when creating the EditWindowVerifyDataSummary.
I have attached a part program and .bmp file showing the issue. Open FLAT1 flatness dimension in summary mode to see it.
<<END>>
<< Changes made by Yanhua Huang (Field Changes) -- 05/19/09 16:16:59>>
Action: Wade Burton to Yanhua Huang
<<END>>
<< Changes made by Neil Kay -- 07/23/08 10:51:37>>
Action: David Petrizze to Wade Burton
<<END>>
<< Paola Pallo -- 03/02/07 12:23:53>>
Fixed.
For review:
v43\dimens\dim_flat.cpp
<<END>>
<< Changes made by Paola Pallo -- 03/02/07 12:23:31>>
Action: Paola Pallo to David Petrizze, Status: OPEN to REVIEW
<<END>>
<< Changes made by Tim Wernicke -- 02/27/07 13:20:13>>
Priority: to High
<<END>> |
|