几何尺寸与公差论坛

 找回密码
 注册

QQ登录

只需一步,快速开始

查看: 248|回复: 0

PR 247925 System takes over half an hour to calculate Profile of 40 to 60 poi

[复制链接]
发表于 2009-8-16 20:06:25 | 显示全部楼层 |阅读模式
System takes over half an hour to calculate Profile of 40 to 60 points
<< Jeff Master  --  05/02/07  14:23:09>>
Customer has 2 ?1000 arms, 1 - 3000i, and 1 ?Infinite.  They are running PC-DMIS 4.1 With *.dll 2.475
When performing profile checks the system may take over half an hour to finish a calculation of 40 points or the system will crash.  Customer sets up an alignment then does a distance scan selecting the CAD entities.  They also have point mode enabled.  Then they take all the points and build a Set.  Then do the Form only profile on the set.  In the Status window you will see it cycle through all the points over and over until it finishes or crashes.  It also did this when performing a Best fit alignment a few times.
<<END>>

<< Changes made by Jeff Master -- 07/26/07  14:56:29>>
Status:  RESOLVED to CLOSED
<<END>>
<< Don Turcotte  --  06/05/07  15:11:31>>
Looking at your debug file I see that you have
  Option/IterateAndRePierceCadOnBestFitAlignments
set to TRUE.  
You don't have
   Option/IterateAndRePierceCadOnBestFitAlignmentsMaxLoops
set so this will default to 100.
You also don't have
   Option/IterateAndRePierceCadOnBestFitAlignmentsTolerance InMM
set so this will default to 20mm.  This should be something slightly larger than the tip diameter (e.g., tip diameter + 0.1mm).
Repiercing cad will greatly slow down form only profile since this uses a 2d bestfit (also explains issues with the 2d bestfit alignment).  We would recommend turning off the repierce cad option unless you really intended to do that.  You will have to use the SettingsEditor.exe to do this.  If you really want to repierce cad, then you should set the MaxLoops and ToleranceInMM options to realistic values.
<<END>>
<< Changes made by Don Turcotte -- 06/05/07  15:11:41>>
Action:  Don Turcotte to Jeff Master, Status:  OPEN to RESOLVED
<<END>>
<< Don Turcotte  --  05/24/07  16:53:26>>
I have received the files and uploaded the zip to the server.  I will look at this issue next week.  I have discovered several problems that are causing slowdowns while working on 248101.
<<END>>
<< Jeff Master  --  05/22/07  08:41:52>>
Sent files 5-22-07
<<END>>
<< Don Turcotte  --  05/11/07  10:51:43>>
Sent e-mail to Jeff :
"Jeff,
I still don? see your files for this PR on the server.  Could you just e-mail them to me and I will put them on the server.
"
<<END>>
<< Changes made by Tim Wernicke -- 05/09/07  14:19:33>>
Status:  MOREINFO to OPEN
<<END>>
<< Jeff Master  --  05/09/07  14:11:27>>
Attached files again
<<END>>
<< Changes made by Jeff Master -- 05/09/07  14:11:30>>
Action:  Jeff Master to Don Turcotte
<<END>>
<< Don Turcotte  --  05/08/07  14:34:55>>
The debug files are not on the server.  In fact, I looked directly at the server array1 drive and there are no files for this PR (there is not even a directory entry for this PR), even though the File Attachments button shows that there are two debug files.
Could you please re-attach your files.  It would be helpful if you could also get the customer's program since often problems are related to a particular sequence of commands in the customer's program.
<<END>>
<< Changes made by Don Turcotte -- 05/08/07  14:35:44>>
Action:  Don Turcotte to Jeff Master, Status:  OPEN to MOREINFO
<<END>>
<< Changes made by Tim Wernicke -- 05/02/07  15:29:47>>
Action:  Tim Wernicke to Don Turcotte, Status:  MOREINFO to OPEN
<<END>>
<< Jeff Master  --  05/02/07  15:15:30>>
Do not have a program added debug file.
<<END>>
<< Changes made by Jeff Master -- 05/02/07  15:15:36>>
Action:  Jeff Master to Tim Wernicke
<<END>>
<< Tim Wernicke  --  05/02/07  14:30:01>>
Hi Jeff, please attach the part program.
<<END>>
<< Changes made by Tim Wernicke -- 05/02/07  14:30:07>>
Action:  Tim Wernicke to Jeff Master, Assigned:   to Don Turcotte, Status:  OPEN to MOREINFO, Priority:  to Critical
<<END>>
您需要登录后才可以回帖 登录 | 注册

本版积分规则

QQ|Archiver|小黑屋|几何尺寸与公差论坛

GMT+8, 2026-4-29 09:10 , Processed in 0.050817 second(s), 19 queries .

Powered by Discuz! X3.4 Licensed

© 2001-2023 Discuz! Team.

快速回复 返回顶部 返回列表