<< Domenico Varacalli  --  03/05/08  13:53:44>>
The sequence of the operations for which the problem happens is not always the same.
In general this happens when I open a window of an already existing feature (FCF or Autofeature).
In this example the error happens when I press the button OK of confirmation, and PC-DMIS it performs the updating  of the whole program,
visualizing the blue scroll bar (1.GIF). However quickly working with the FCF and the Autofeatures, this type of crash appears with an elevated frequency.
In the ZIP file there are the debug data and the video.
<<END>>
 楼主| 发表于 2008-5-26 14:39:31 | 显示全部楼层

回复: PR 253655 Application Error during the use of the FCF and Autofeatures.

<< Don Turcotte  --  03/17/08  16:56:25>>
I fixed two issues:
1.  F9 on feature, then OK, F9 on FCF, then crash.
2.  Delete Simultaneous evaluation, then F9 on feature, then crash.

The issue of the two messages for "Do you want to update nominals of related dimensions?" when F9 on feature, then OK, has been filed as PR#253846.

Fixed in V43B and V44B.

Files inserted to server
------------------------
V43B\DIMENS\FEATCTRLFRM.CPP
V43B\INCLUDE\FEATCTRLFRM.H

V44B\DIMENS\FEATCTRLFRM.CPP
V44B\INCLUDE\FEATCTRLFRM.H

<<END>>